Results 1 to 16 of 16

Thread: Up to 2 degrees

  1. #1
    Bad address email on file
    Join Date
    Feb 2019
    Location
    Washington State
    Occupation
    Optical Laboratory Technician
    Posts
    16

    Up to 2 degrees

    ANSI standards say i can have up to 2 degrees of tilt in my multifocal lenses. how do i go about measuring those 2 degrees? ive looked all over for a layout chart and can find nothing on the web.

  2. #2
    Rising Star
    Join Date
    Sep 2015
    Location
    Somerville, MA
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    52
    Quote Originally Posted by Dooky Shoes View Post
    ANSI standards say i can have up to 2 degrees of tilt in my multifocal lenses. how do i go about measuring those 2 degrees? ive looked all over for a layout chart and can find nothing on the web.
    Just search for pantoscopic tilt instead.

    Here's some info: https://www.odob.health.nz/wp-conten...f-D-Wilson.pdf
    Vitor Pamplona
    CEO, EyeNetra Inc

    Come check our autorefractor, lensometer and phoropter.

  3. #3
    Bad address email on file
    Join Date
    Feb 2019
    Location
    Washington State
    Occupation
    Optical Laboratory Technician
    Posts
    16
    Quote Originally Posted by vfpamp View Post
    Just search for pantoscopic tilt instead.

    Here's some info: https://www.odob.health.nz/wp-conten...f-D-Wilson.pdf
    that has nothing to do with what i was asking. i'll give you an example.

    Lets say i have a freshly edged and mounted pair of bifocals. everything checks out on the lensometer. all my axis's and powers are spot on. BUT! the bi-focal segment is crooked by an undetermined amount. how would i go about measuring how far off of the 180 that segment is? or am i just coming at this all wrong?

  4. #4
    Rising Star
    Join Date
    Sep 2015
    Location
    Somerville, MA
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    52
    Quote Originally Posted by Dooky Shoes View Post
    that has nothing to do with what i was asking. i'll give you an example.

    Lets say i have a freshly edged and mounted pair of bifocals. everything checks out on the lensometer. all my axis's and powers are spot on. BUT! the bi-focal segment is crooked by an undetermined amount. how would i go about measuring how far off of the 180 that segment is? or am i just coming at this all wrong?
    Ohh... But if the axis checks out for the distance part, doesn't that say that you mounted the bifocal segment of the lens correctly? If you rotate one, you have to rotate the other. The only option is a manufacturing defect on the bifocal blank (either on the axis of astigmatism or on the bifocal segment).
    Vitor Pamplona
    CEO, EyeNetra Inc

    Come check our autorefractor, lensometer and phoropter.

  5. #5
    OptiBoard Novice
    Join Date
    Jan 2020
    Location
    United States
    Occupation
    Optical Laboratory Technician
    Posts
    7
    Can you turn the lens in the frame at all? In this scenario I would try getting the bifocal seg straight and then see if the cyl is within tolerance. If you can't simultaneously get the bifocal and cyl to be correct, that sounds like a manufacturing/surfacing error, as vfpamp said.

  6. #6
    OptiBoardaholic OptiBoard Bronze Supporter
    Join Date
    Jan 2017
    Location
    USA
    Occupation
    Dispensing Optician
    Posts
    216
    If you needed a precise measurement for determining how crooked a seg is, I'd just take it over to the lensometer and mark the 180 line on the mounted lenses and use a protractor to find the angle of the seg.

    I usually just hold a PD ruler across the top of both segs and make sure they are parallel to eachother. If they are way off to the point of needing a remake, it's usually plainly obvious without needing to use a tool or measure it.

  7. #7
    Master OptiBoarder
    Join Date
    Feb 2016
    Location
    usa
    Occupation
    Dispensing Optician
    Posts
    989
    Quote Originally Posted by vfpamp View Post
    Ohh... But if the axis checks out for the distance part, doesn't that say that you mounted the bifocal segment of the lens correctly? If you rotate one, you have to rotate the other. The only option is a manufacturing defect on the bifocal blank (either on the axis of astigmatism or on the bifocal segment).
    No. ANSI Z80.1-2015 has a tolerance of 14° for up to a .25 cylinder, 7° for .25-.50 cyl etc. So obviously the axis of the lens doesn't need to match the axis of the seg line to be withing ANSI suggested tolerance.

  8. #8
    Bad address email on file
    Join Date
    Feb 2019
    Location
    Washington State
    Occupation
    Optical Laboratory Technician
    Posts
    16
    Quote Originally Posted by Kwill212 View Post
    No. ANSI Z80.1-2015 has a tolerance of 14° for up to a .25 cylinder, 7° for .25-.50 cyl etc. So obviously the axis of the lens doesn't need to match the axis of the seg line to be withing ANSI suggested tolerance.
    exactly! the place where this comes into play the most is with larger cyls. but i think that its more of a non issue than i was thinking at first.

  9. #9
    Rising Star
    Join Date
    Sep 2015
    Location
    Somerville, MA
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    52
    Quote Originally Posted by Kwill212 View Post
    No. ANSI Z80.1-2015 has a tolerance of 14° for up to a .25 cylinder, 7° for .25-.50 cyl etc. So obviously the axis of the lens doesn't need to match the axis of the seg line to be withing ANSI suggested tolerance.
    oh I see. When you said "my axis's and powers are spot on" I assumed dead on, not only within the ANSI Limits.


    Vitor Pamplona
    CEO, EyeNetra Inc

    Come check our autorefractor, lensometer and phoropter.

  10. #10
    Master OptiBoarder OptiBoard Silver Supporter Barry Santini's Avatar
    Join Date
    May 2000
    Location
    Seaford, NY USA
    Occupation
    Dispensing Optician
    Posts
    6,008
    All these are manufacturing/lab standards, not dispensing standards.

    B

  11. #11
    OptiBoard Professional Mauro.Airoldi's Avatar
    Join Date
    May 2007
    Location
    Bologna Italy
    Occupation
    Lens Manufacturer
    Posts
    161
    The reference in Europe is EN ISO 2187 and give you all the information

  12. #12
    Bad address email on file
    Join Date
    Feb 2019
    Location
    Washington State
    Occupation
    Optical Laboratory Technician
    Posts
    16
    Quote Originally Posted by Barry Santini View Post
    All these are manufacturing/lab standards, not dispensing standards.

    B
    i work in a lab

  13. #13
    Master OptiBoarder OptiBoard Silver Supporter ak47's Avatar
    Join Date
    Nov 2010
    Location
    Southwest US
    Occupation
    Dispensing Optician
    Posts
    567
    I don't think anyone has answered the OP's question, so I'll take a stab. ANSI HORIZONTAL SEGMENT TILT < 2 DEGREES CROOKED......I asked and answered this question of myself a while back with trigonometry knowing that the dots are 34 mm apart.....and the answer is slightly over a 1 mm vertical height difference between the nasal dot and the temporal dot. I used tangent (2 degrees = 0.0349066 radians) = x / 34...therefore x = 1.2 mm

  14. #14
    Bad address email on file
    Join Date
    Feb 2019
    Location
    Washington State
    Occupation
    Optical Laboratory Technician
    Posts
    16
    Quote Originally Posted by ak47 View Post
    I don't think anyone has answered the OP's question, so I'll take a stab. ANSI HORIZONTAL SEGMENT TILT < 2 DEGREES CROOKED......I asked and answered this question of myself a while back with trigonometry knowing that the dots are 34 mm apart.....and the answer is slightly over a 1 mm vertical height difference between the nasal dot and the temporal dot. I used tangent (2 degrees = 0.0349066 radians) = x / 34...therefore x = 1.2 mm
    that would work for a progressive, but how would you do a bifocal?

  15. #15
    Master OptiBoarder OptiBoard Silver Supporter ak47's Avatar
    Join Date
    Nov 2010
    Location
    Southwest US
    Occupation
    Dispensing Optician
    Posts
    567
    The line at the top of a FT28 is 26 mm long....so the difference in vertical height to be within 2 degrees tolerance from nasal side of seg line to temporal side would be (26/34) x 1.2 mm = 0.9 mm

  16. #16
    Master OptiBoarder AngeHamm's Avatar
    Join Date
    Apr 2010
    Location
    Richmond, VA
    Occupation
    Optical Retail
    Posts
    2,370
    A protractor is your easiest bet.
    I'm Andrew Hamm and I approve this message.

Thread Information

Users Browsing this Thread

There are currently 1 users browsing this thread. (0 members and 1 guests)

Similar Threads

  1. Online Associates Degrees?
    By ALASKA in forum Professional and Educational Organizations Discussion Forum
    Replies: 3
    Last Post: 03-25-2011, 12:44 PM
  2. Diopters and Degrees
    By OPTIDONN in forum Ophthalmic Optics
    Replies: 23
    Last Post: 12-23-2005, 08:58 AM
  3. Opticians with AS degrees
    By Laurie in forum General Optics and Eyecare Discussion Forum
    Replies: 9
    Last Post: 04-03-2001, 03:33 PM
  4. 360 degrees
    By harry a saake in forum General Optics and Eyecare Discussion Forum
    Replies: 7
    Last Post: 11-27-2000, 11:02 PM