Essilor Kappa - tracer problem

Collapse
X
 
  • Filter
  • Time
  • Show
Clear All
new posts

  • Essilor Kappa - tracer problem

    Hello,
    I have a problem with tracer essilor kappa. The slider falls during scanning. Has anyone tried to deal with the problem by themselves? I know this is a glitch for the service technician but maybe someone has dealt with this without the help of the essilor service?
    I am also looking for the manual of this equipment, does anyone have such?

  • #2
    What do you mean by "slider"? is it the tracing pin? If so, I find that if it's a wrap frame I have to push down on the temples during the trace to sort of flatten the front momentarily. Also, sometimes there's a flaw, maybe a drop of solder in the frame bevel or a screw is loose in which case you may be able to fix it or you may need to dot up the sample lens and trace it.

    Comment


    • #3
      It just occurred to me you may mean the shelf that you set the frame into which then clamps down on the frame. There are a few frames that those clamps won't hold and I have hold that "slider" really hard so it won't slip.

      Comment


      • #4
        The user's guide is 234 pages long; no one would want to do all that scanning. Call Essilor 855-393-4647 option 3 to order one.

        Comment


        • #5
          yes, "slider" is it "tracing pin" :)
          replacing the entire item (as in the picture) helped. The bearings have worn out.
          Pressing the frame so that the pin does not fall is a solution but it distorts the 3D measurement.