Thread: ANSI Z80.1 Prism inspection - question

  1. #1
    OptiBoard Apprentice
    Join Date
    Apr 2015
    Location
    Belgique
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    12

    ANSI Z80.1 Prism inspection - question

    Hi everyone,

    I have a question regarding the prism inspection as described by the ANSI Z80.1 (2010) standard.
    Indeed, it seems to me that there is a contradiction between two paragraphs and I would like to know your opinion.

    Here are the two excerpts :
    5.1.4 Prism Reference Point Location and Prismatic Power
    The prism reference point shall not be more than 1.0 mm away from its specified
    position in any direction. In addition, the prismatic power measured at the prism
    reference point shall not exceed 0.33cm/m. [...] Measurement shall be done using the method in 8.4.
    8.4 Prismatic Power Measurement
    1. Position the lens on the focimeter stage with the prism reference point centered in
    front of the lens stop of the focimeter. [...]
    2. Measure the prism at the PRP. Proceed to Step 3 only if the prism error exceeds
    0.33cm/m at the prism reference point.
    3. Locate and mark the position at which the prismatic requirements of the
    prescription are met. In the absence of prescribed prism this point corresponds
    with the optical center. This point shall be within 1.0 mm of the prism reference
    point as specified in 5.1.4.



    In other words, for a lens without a prescribed prism,
    • paragraph 5.1.4 says : "the prism at the PRP should be lower than 0.33cm/m AND the PRP should be at a maximum distance of 1mm regarding the optical center.
    • paragaph 8.4 says : "the prism at the PRP should be lower than 0.33cm/m OR the PRP should be at a maximum distance of 1mm regarding the the optical center


    The "AND"/"OR" confusion makes a big difference.
    For example, if we consider the case of a spherical lens and translate the tolerance on the decentration on the prism value based on the Prentice rule, we can get a single value for the tolerance on the prismatic power. Using the "AND" or "OR" leads to a very different tolerance as illustrated here below :

    Click image for larger version. 

Name:	psmTol.png 
Views:	80 
Size:	17.0 KB 
ID:	11904

    I believe paragraph 5.1.4 is wrong since the tolerance it implies would be impossible to apply (for a plano lens, the tolerance is equal to zero!, and for a +10D, a shift of 0.3mm would make the lens fail).

    Do you agree with me?

    What rule do you use to inspect the prism based on the ANSI standard?

    Thanks in advance!

    Thomas
    Last edited by thomasz; 04-27-2015 at 11:03 AM.

  2. #2
    Banned
    Join Date
    Jul 2010
    Location
    St. Cloud, Minnesota
    Occupation
    Ophthalmic Technician
    Posts
    3,089
    I read the two paragraphs as identical. 8.4.3 gives the method of determining the amount of error and states that it "SHALL BE" within 1.0 mm.

    I believe that there are exceptions for plano power prismatic lenses elsewhere in the standard.

  3. #3
    Eyes eastward... Uilleann's Avatar
    Join Date
    Jan 2007
    Location
    Utah
    Occupation
    Dispensing Optician
    Posts
    3,246
    Sounds like an ABO test writer had a hand writing the ANSI standards as well.

  4. #4
    Master OptiBoarder lensgrinder's Avatar
    Join Date
    Jul 2005
    Location
    Raleigh, NC
    Occupation
    Lens Manufacturer
    Posts
    506
    5.1.4 and 8.4 are general tolerances. They are to be used for unmounted lenses. Once you mount the lenses this will change the PRP locations. This is why the tolerances are different for mounted pairs.
    In regards to plano, 3.21.4 states "Unless otherwise specified, the prism reference point is assumed to be at thefollowing geometric locations on the lens. For uncut single vision lenses, this point is the geometric center of the lens." So if you have 0.33 D at the geometric center of a plano then it will pass.

  5. #5
    Banned
    Join Date
    Jul 2010
    Location
    St. Cloud, Minnesota
    Occupation
    Ophthalmic Technician
    Posts
    3,089
    ^^+1000

  6. #6
    OptiBoard Apprentice
    Join Date
    Apr 2015
    Location
    Belgique
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    12
    @ lensgrinder
    Thanks for your reply.
    The distinction between uncut and mounted is not relevant to me since 5.1.4 and 8.4 are general paragraphs.
    I also believe that a plano with a prism < 0.33cm/m at the PRP would need to pass.
    But I also believe that this is in contradiction with §5.1.4. Please read it again.

  7. #7
    OptiBoard Apprentice
    Join Date
    Apr 2015
    Location
    Belgique
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    12
    @MikeAurelius.
    Please read them again then.
    To, me the logic is clearly different.
    If A and B are two conditions, (A OR B) is different from (A AND B)!

  8. #8
    Banned
    Join Date
    Jul 2010
    Location
    St. Cloud, Minnesota
    Occupation
    Ophthalmic Technician
    Posts
    3,089
    I believe you are overthinking it. As lensgrinder states, for uncut lenses SV it doesn't matter, as long as the finished lens cuts out correctly with the prescribed prism.

    From 5.1.4:

    The prism reference point shall not be more than 1.0 mm away from its specified position in any direction. In addition, the prismatic power measured at the prism
    reference point shall not exceed 0.33cm/m
    This is clearly (A AND B), note the "In Addition" phrase.

  9. #9
    Master OptiBoarder MakeOptics's Avatar
    Join Date
    Aug 2011
    Location
    none
    Occupation
    Other Eyecare-Related Field
    Posts
    1,327
    Quote Originally Posted by thomasz View Post
    Hi everyone,

    I have a question regarding the prism inspection as described by the ANSI Z80.1 (2010) standard.
    Indeed, it seems to me that there is a contradiction between two paragraphs and I would like to know your opinion.

    Here are the two excerpts :
    5.1.4 Prism Reference Point Location and Prismatic Power
    The prism reference point shall not be more than 1.0 mm away from its specified
    position in any direction. In addition, the prismatic power measured at the prism
    reference point shall not exceed 0.33cm/m. [...] Measurement shall be done using the method in 8.4.
    8.4 Prismatic Power Measurement
    1. Position the lens on the focimeter stage with the prism reference point centered in
    front of the lens stop of the focimeter. [...]
    2. Measure the prism at the PRP. Proceed to Step 3 only if the prism error exceeds
    0.33cm/m at the prism reference point.
    3. Locate and mark the position at which the prismatic requirements of the
    prescription are met. In the absence of prescribed prism this point corresponds
    with the optical center. This point shall be within 1.0 mm of the prism reference
    point as specified in 5.1.4.



    In other words, for a lens without a prescribed prism,
    • paragraph 5.1.4 says : "the prism at the PRP should be lower than 0.33cm/m AND the PRP should be at a maximum distance of 1mm regarding the optical center.
    • paragaph 8.4 says : "the prism at the PRP should be lower than 0.33cm/m OR the PRP should be at a maximum distance of 1mm regarding the the optical center


    The "AND"/"OR" confusion makes a big difference, as illustrated below :

    Click image for larger version. 

Name:	psmTol.png 
Views:	80 
Size:	17.0 KB 
ID:	11904

    I believe paragraph 5.1.4 is wrong since the tolerance it implies would be impossible to apply (for a plano lens, the tolerance is equal to zero!, and for a +10D, a shift of 0.3mm would make the lens fail).

    Do you agree with me?

    What rule do you use to inspect the prism based on the ANSI standard?

    Thanks in advance!

    Thomas
    Looking at your graph you forgot a very important part of the equation, the domain.

    0.33PD is only applied to prescription power 0 to <= 3.37 and the 1mm or less is applied to the >3.37 powers.
    http://www.opticians.cc

    Creator of the industries 1st HTML5 Browser based tracer software.
    Creator of the industries 1st Mac tracer software.
    Creator of the industries 1st Linux tracer software.

  10. #10
    OptiBoard Apprentice
    Join Date
    Apr 2015
    Location
    Belgique
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    12
    @MikeAurelius.
    In my situation, it is very important to understand exactly what the standard means.
    I agree with you regarding §5.1.4 but, if you read §8.4 carefully, you will realize it is a "OR" in that case.
    Since a "AND" would make no sense in practice, I believe §8.4 is right and §5.1.4 is wrong.
    What do you think?

  11. #11
    OptiBoard Apprentice
    Join Date
    Apr 2015
    Location
    Belgique
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    12
    @MakeOptics
    Sorry if I wasn't clear. Actually, to be able to compare the two conditions under the same graphs, I consider the case of a spherical lens.
    For such a lens, the relation between the prism and the decentration is given by the Prentice rule : PRISM[cm/m] = 0.1*POWER[D]*DECENTRATION[mm].
    So I didn't forget the domain actually ;-)
    In that case, you can see how the two conditions combined with a "OR" or "AND" and translated into a prism tolerance can lead to a totally different inspection result.

  12. #12
    Banned
    Join Date
    Jul 2010
    Location
    St. Cloud, Minnesota
    Occupation
    Ophthalmic Technician
    Posts
    3,089
    Quote Originally Posted by thomasz View Post
    @MikeAurelius.
    In my situation, it is very important to understand exactly what the standard means.
    I agree with you regarding §5.1.4 but, if you read §8.4 carefully, you will realize it is a "OR" in that case.
    Since a "AND" would make no sense in practice, I believe §8.4 is right and §5.1.4 is wrong.
    What do you think?
    I think the mistake you've made is in thinking that both paragraphs are somehow equivalent, when they are not. When working with standards of any kind there are usually at least two parts, the description of the specification, IE: what it is you are measuring and what the allowable tolerances are, paragraph 5.1.4 gives you that. The other part is the method by which you test the item to see if it meets the specification of the standard, and that's whats in 8.4.

    5.1.4 is the SPECIFICATION, 8.4 is the METHOD.

    Two parts, one result, pass or fail.

  13. #13
    OptiBoard Apprentice
    Join Date
    Apr 2015
    Location
    Belgique
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    12
    MikeAurelius, I agree with you on the two parts, one giving the concept and one giving the procedure to apply this concept.
    However, I really don't agree with you in that case since the concept is clearly different between the two parts.
    You probably understand since you actually changed your mind. 5.1.4 stated (A and B), which you believed was right... 8.4 says (A or B), which you seem to believe now...
    So you see how confusing it is :-)?

  14. #14
    Banned
    Join Date
    Jul 2010
    Location
    St. Cloud, Minnesota
    Occupation
    Ophthalmic Technician
    Posts
    3,089
    Quote Originally Posted by thomasz View Post
    MikeAurelius, I agree with you on the two parts, one giving the concept and one giving the procedure to apply this concept.
    However, I really don't agree with you in that case since the concept is clearly different between the two parts.
    You probably understand since you actually changed your mind. 5.1.4 stated (A and B), which you believed was right... 8.4 says (A or B), which you seem to believe now...
    So you see how confusing it is :-)?
    I haven't changed my mind. The wording is slightly different, however the intent is very clear to me (btw, I read specifications all the time, not just in the optical industry).

    8.4 says to do this:

    1. Locate lens on the lensometer at the PRP. (assumes you've already marked the PRP)
    2. Read the prismatic power. If the difference between Rx and read power is greater than .33 prism diopters, proceed to step 3.
    3. Move the lens in the lensometer until it matches Rx. Mark the lens. Remove lens and measure distance between PRP and actual prismatic location. If the measurement is 1 mm or less, the lens is good.

    Plus or minus 1.00 mm in positioning error is the tolerance.

    There's nothing to confuse here, its very straight forward.

  15. #15
    OptiBoard Apprentice
    Join Date
    Apr 2015
    Location
    Belgique
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    12
    That's a bit funny Mike :-)

    Let's say that I agree with you, it is (A or B)!
    Either the prism is lower than 0.33D or the distance from the PRP to the point where the prescribed prism is met is lower than 1mm.
    Just not 0.33D and distance lower than 1mm.

    If you consider a non prismatic lens with a very low power, you can have a prism of 0.25cm/m at the PRP and an optical center that is 10mm from there.

    So I guess we agree 8.4 is right :-)

  16. #16
    Banned
    Join Date
    Jul 2010
    Location
    St. Cloud, Minnesota
    Occupation
    Ophthalmic Technician
    Posts
    3,089
    In a lens with prism, there is no OC, just PRP.

    Erg...that came out wrong...

    S/B: When edging a lens with prism, the only thing (especially in SV) that matters is the PRP.
    Last edited by MikeAurelius; 04-27-2015 at 07:49 PM.

  17. #17
    Master OptiBoarder lensgrinder's Avatar
    Join Date
    Jul 2005
    Location
    Raleigh, NC
    Occupation
    Lens Manufacturer
    Posts
    506
    Quote Originally Posted by thomasz View Post
    @ lensgrinder
    Thanks for your reply.
    The distinction between uncut and mounted is not relevant to me since 5.1.4 and 8.4 are general paragraphs.
    I also believe that a plano with a prism < 0.33cm/m at the PRP would need to pass.
    But I also believe that this is in contradiction with §5.1.4. Please read it again.
    It is relevant to ANSI. This is why they have general tolerances for uncut and edged versus mounted. Also note that the general tolerances do not reference power in regards to prism, which is another reason they are general.

    8.4
    If you have an edged lens that is 50 mm round with 5 mm of decentration and no prescribed prism then your PRP should be located at 30 mm from the temporal edge. If, at this point you have 0.33 then it is OK. If you notice more prism than 0.33 , you will move the lens until it has no prism. If this movement is greater than 1 mm then it fails.

    5.1.4
    You need to fall within 1mm of the specified PRP(30 mm in the above case) at the PRP you can have 0.33


    The way to measure this is to place the lens in the
    focimeter at the requested PRP and if it has prism greater than 0.33 then move it and if it is not within 1mm of the desired PRP then it fails.

    They are saying the same thing.

  18. #18
    OptiWizard
    Join Date
    Apr 2014
    Location
    California
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    382
    Quote Originally Posted by thomasz View Post
    Hi everyone,

    Here are the two excerpts :
    5.1.4 Prism Reference Point Location and Prismatic Power
    The prism reference point shall not be more than 1.0 mm away from its specified
    position in any direction. In addition, the prismatic power measured at the prism
    reference point shall not exceed 0.33cm/m. [...] Measurement shall be done using the method in 8.4.
    8.4 Prismatic Power Measurement
    1. Position the lens on the focimeter stage with the prism reference point centered in
    front of the lens stop of the focimeter. [...]
    2. Measure the prism at the PRP. Proceed to Step 3 only if the prism error exceeds
    0.33cm/m at the prism reference point.
    3. Locate and mark the position at which the prismatic requirements of the
    prescription are met. In the absence of prescribed prism this point corresponds
    with the optical center. This point shall be within 1.0 mm of the prism reference
    point as specified in 5.1.4.



    In other words, for a lens without a prescribed prism,
    • paragraph 5.1.4 says : "the prism at the PRP should be lower than 0.33cm/m AND the PRP should be at a maximum distance of 1mm regarding the optical center.
    • paragaph 8.4 says : "the prism at the PRP should be lower than 0.33cm/m OR the PRP should be at a maximum distance of 1mm regarding the the optical center

    Don't let your English skills hinder your logic.

  19. #19
    OptiBoard Apprentice
    Join Date
    Apr 2015
    Location
    Belgique
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    12
    Hi lensgrinder and ml43,

    Thanks for taking the time to respond!
    I'm sorry but I don't follow you guys :-)

    I understand the discussion for decentered SV can be more complex.
    However, the discussion is also valid for a progressive lens with a PRP at the center of the engravings.

    Let's turn this in mathematical terms and let's consider the case of a progressive lens where the prescribed prism equal to zero.
    Let's define PRP = center of the engravings and OC = point where the prism equal zero.

    A : PRISM(PRP)<0.33cm/m
    B : distance(PRP,OC) < 1mm

    To me, 5.1.4 says lens passes if (A and B).
    And 8.4 says : (A or((not A) and B)) which, as you can verify, is exactly the same logic as (A or B).

    I'm sorry to insist.
    Maybe I'm the only one to believe there is a problem :-)

  20. #20
    OptiBoard Apprentice
    Join Date
    Apr 2015
    Location
    Belgique
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    12
    @MikeAurelius
    Sorry if I wasn't clear. I'm considering a centered SV lens without any prescribed prism but with a power close to zero.
    Let's say SPH=0.25D, PRVM=0.

    During the inspection, the measured prism at the PRP=geometrical center could be
    INSPRVM=0.25cm/m. -> this prism is due to manufacturing incertitude.

    It passes the A condition (lower than 0.33cm/m).
    However, the optical center where the prism is equal to zero is at a distance of 10mm (prentice rule).
    So B condition will fail.

    Based on 8.4, the lens should pass. But 5.1.4 tends to say the opposite.

  21. #21
    OptiWizard
    Join Date
    Apr 2014
    Location
    California
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    382
    Quote Originally Posted by thomasz View Post
    @MikeAurelius
    Sorry if I wasn't clear. I'm considering a centered SV lens without any prescribed prism but with a power close to zero.
    Let's say SPH=0.25D, PRVM=0.

    During the inspection, the measured prism at the PRP=geometrical center could be
    INSPRVM=0.25cm/m. -> this prism is due to manufacturing incertitude.

    It passes the A condition (lower than 0.33cm/m).
    However, the optical center where the prism is equal to zero is at a distance of 10mm (prentice rule).
    So B condition will fail.

    Based on 8.4, the lens should pass. But 5.1.4 tends to say the opposite.
    Your logic is being corrupted by your reading comprehension and order of operations.

    5.1.4 and 8.4 are complementary to each other, and are worded quite clever.
    If you read carefully, what it's saying is:

    Unmounted Prism and PRP ≥ 0.00 D, ≤ ±3.37 D
    > ±3.37 D
    0.33 Δ
    1.0 mm




    In this case the lens would pass both 5.1.4 and 8.4
    remember, in a SV SPH 0.25D lens, the PRP is the GC, not the OC.

  22. #22
    Banned
    Join Date
    Jul 2010
    Location
    St. Cloud, Minnesota
    Occupation
    Ophthalmic Technician
    Posts
    3,089
    Please do not take offense at this question, but is English your first language or a subsequent language?

    This standard has stood *as-is* for the better part of 30 years and probably even longer. *IF* there had been an error in logic, it would have been caught by the hundreds of thousands of people who use it every single day as it is written.

    If you want to believe it's a mistake, then write to ANSI and file a brief with them detailing what you believe the error to be. Send a copy to the VCA (Vision Council of America), as they are the ones most likely to respond.

    We've explained to you that it's your logic that is in error, but you can't seem to accept that.

  23. #23
    OptiBoard Apprentice
    Join Date
    Apr 2015
    Location
    Belgique
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    12
    English is not my first language as you probably noticed ;-)
    I got your point. Maybe it is a matter of language...
    I will address my concern to the Vision Council.
    Thank you!

  24. #24
    Master OptiBoarder lensgrinder's Avatar
    Join Date
    Jul 2005
    Location
    Raleigh, NC
    Occupation
    Lens Manufacturer
    Posts
    506
    Quote Originally Posted by thomasz View Post
    Let's turn this in mathematical terms and let's consider the case of a progressive lens where the prescribed prism equal to zero.
    Let's define PRP = center of the engravings and OC = point where the prism equal zero.

    A : PRISM(PRP)<0.33cm/m
    B : distance(PRP,OC) < 1mm

    To me, 5.1.4 says lens passes if (A and B).
    And 8.4 says : (A or((not A) and B)) which, as you can verify, is exactly the same logic as (A or B).
    If a lens has zero prism diopters then it is still referred to as the PRP, you would be referencing zero prism diopters. There is no OC.
    In a progressive you have a known PRP, so if you check it at that location and it is <0.33 then it will pass > 0.33 it fails. There is not within 1 mm because you know where the PRP is.

    Quote Originally Posted by thomasz View Post
    I'm considering a centered SV lens without any prescribed prism but with a power close to zero.
    Let's say SPH=0.25D, PRVM=0.

    During the inspection, the measured prism at the PRP=geometrical center could be
    INSPRVM=0.25cm/m. -> this prism is due to manufacturing incertitude.

    It passes the A condition (lower than 0.33cm/m).
    However, the optical center where the prism is equal to zero is at a distance of 10mm (prentice rule).
    So B condition will fail.

    Based on 8.4, the lens should pass. But 5.1.4 tends to say the opposite.
    If the prism is 0.25 at the PRP then it will pass.

  25. #25
    OptiBoard Apprentice
    Join Date
    Apr 2015
    Location
    Belgique
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    12
    Hi lensgrinder,

    I don't agree with you on both points :

    - For uncut SV lenses without a prescribed prism, the PRP is the geometrical center and can be different from the OC where the prism is equal to zero, since the manufacturing is not perfect. There is an OC and there is a PRP and those points can be different.

    - For a progressive lens, based on §8.4, the same procedure applies. Not understanding that is a serious flaw in your understanding of the optical principles behind the procedure. If you have a +15.00D lens, an error of positioning of the PRP at the focimeter of 0.2mm (which is low and can happen all the time with manual positioning) would lead to an error on the prism of 0.30cm/m. Therefore, the tolerance on the prism will easily fails, while the tolerance on the distance between the PRP and the OC (or point where the prescription prism is met) can pass, so the lens will pass.

Thread Information

Users Browsing this Thread

There are currently 1 users browsing this thread. (0 members and 1 guests)

Similar Threads

  1. Question on ANSI Standards
    By SailorEd in forum General Optics and Eyecare Discussion Forum
    Replies: 8
    Last Post: 03-06-2012, 05:39 PM
  2. Ansi Horizontal Prism Revisted
    By Aarlan in forum General Optics and Eyecare Discussion Forum
    Replies: 17
    Last Post: 06-02-2008, 02:42 PM
  3. ANSI standards...question
    By gralena79 in forum General Optics and Eyecare Discussion Forum
    Replies: 1
    Last Post: 05-09-2008, 08:48 PM
  4. Question about ANSI standards - specifically PD
    By co_mac in forum Ophthalmic Optics
    Replies: 7
    Last Post: 01-17-2008, 01:02 PM
  5. ANSI question for Darryl
    By OPTIDONN in forum General Optics and Eyecare Discussion Forum
    Replies: 1
    Last Post: 06-18-2005, 11:29 PM