    lens meter parametring

    I am using 2 shinipon lens meter in my lab and there is a difference between the 2 one when I mesure By 0.01 step some times the difference is in SPH and some times in Cyl. the difference can be more then 0.1 diopter some times!!
    thanks for your help

    Hamza Mahmoud

    Mail : mahmoud.hamza@optylab.com
    Web : www.optylab.com

  2. #2
    Underemployed Genius Jacqui's Avatar
    Join Date
    Oct 2003
    Location
    Frostbite Falls, Mn.
    Occupation
    Optical Wholesale Lab (other positions)
    Posts
    7,417
    Are they both properly calibrated?? Use the same standards for both machines.

  3. #3
    Rising Star mahmoud.hamza's Avatar
    Join Date
    Oct 2007
    Location
    Tunisia
    Occupation
    Lens Manufacturer
    Posts
    81
    can you help me how to calibrate them?

    Hamza Mahmoud

    Mail : mahmoud.hamza@optylab.com
    Web : www.optylab.com

  4. #4
    Master OptiBoarder
    Join Date
    Oct 2005
    Location
    new york
    Occupation
    Optometrist
    Posts
    3,749
    I'm not familiar with the brand you are using, but how can you tell if the instruement is off by .01?...that's too small an increment...unless you have a digital readout? If this is a basic optical lensometer, and the amount that the instruement is off varies from one read to the next, it is possible that the power wheel is just loose on the shaft. Even .1D is a small amount and is likely just due to operator variability.

  5. #5
    Rising Star mahmoud.hamza's Avatar
    Join Date
    Oct 2007
    Location
    Tunisia
    Occupation
    Lens Manufacturer
    Posts
    81
    I have 2 Shin-nippon lensmeter SLM4000 and SLM5000 one I am using them in the quality control step in my surfacing lab.
    the prolem that the mesurment that I am making with the First SLM4000 is different then the SLM5000 when I use 0.01 step and espacelly in cyl

    an exemple:
    target +3.50
    1st SLM 4000 +3.53 (+0.11)
    SLM 5000 +3.51 (+0.00)


    For your reference the SLM5000 is 3 years old and i did never make a calibration on it because I did not found the calibration set in in Tunisia, the SLM4000 Are new

    Hamza Mahmoud

    Mail : mahmoud.hamza@optylab.com
    Web : www.optylab.com

  6. #6
    ATO Member HarryChiling's Avatar
    Join Date
    Apr 2005
    Location
    Nowhereville
    Occupation
    Other Eyecare-Related Field
    Posts
    7,765
    Quote Originally Posted by mahmoud.hamza View Post
    I have 2 Shin-nippon lensmeter SLM4000 and SLM5000 one I am using them in the quality control step in my surfacing lab.
    the prolem that the mesurment that I am making with the First SLM4000 is different then the SLM5000 when I use 0.01 step and espacelly in cyl

    an exemple:
    target +3.50
    1st SLM 4000 +3.53 (+0.11)
    SLM 5000 +3.51 (+0.00)


    For your reference the SLM5000 is 3 years old and i did never make a calibration on it because I did not found the calibration set in in Tunisia, the SLM4000 Are new

    I have never worked on the model lensometers you own, but do keep in mind that all instruments will have a tolerance, and more than likely you have a tolerance of +/- 0.01D which can account for up to 0.02 difference between measuring the same lens.
    1st* HTML5 Tracer Software
    1st Mac Compatible Tracer Software
    1st Linux Compatible Tracer Software

    *Dave at OptiVision has a web based tracer integration package that's awesome.

  7. #7
    Rising Star mahmoud.hamza's Avatar
    Join Date
    Oct 2007
    Location
    Tunisia
    Occupation
    Lens Manufacturer
    Posts
    81
    but the difference that I have is 0.11!! it is not small difference !

    Hamza Mahmoud

    Mail : mahmoud.hamza@optylab.com
    Web : www.optylab.com

  8. #8
    Master OptiBoarder OptiBoard Silver Supporter Barry Santini's Avatar
    Join Date
    May 2000
    Location
    Seaford, NY USA
    Occupation
    Dispensing Optician
    Posts
    6,009
    Quote Originally Posted by mahmoud.hamza View Post
    but the difference that I have is 0.11!! it is not small difference !
    In my expereince with My Humphrey Lab 360, think "dust/smudges" in the laser path as the culprit for the 0.11 cyl.

    FWIW

    Barry

  9. #9
    Rising Star mahmoud.hamza's Avatar
    Join Date
    Oct 2007
    Location
    Tunisia
    Occupation
    Lens Manufacturer
    Posts
    81
    Barry can you please explain?

    Hamza Mahmoud

    Mail : mahmoud.hamza@optylab.com
    Web : www.optylab.com

  10. #10
    Rising Star mahmoud.hamza's Avatar
    Join Date
    Oct 2007
    Location
    Tunisia
    Occupation
    Lens Manufacturer
    Posts
    81
    also another power :

    sph -2.50
    ==> SLM 4000 => -2.60 (-0.1)
    ==> Essilo => -2.57 (-0.19)
    It make me creazy !!!
    :hammer::hammer::hammer::hammer::hammer::hammer:

    Hamza Mahmoud

    Mail : mahmoud.hamza@optylab.com
    Web : www.optylab.com

  11. #11
    Bad address email on file
    Join Date
    May 2009
    Location
    Prague
    Occupation
    Optical Wholesale Lab (other positions)
    Posts
    5
    We do a calibration on all our lensmeters at least once a year , on the ones in QC we calibrate more often.

  12. #12
    Rising Star mahmoud.hamza's Avatar
    Join Date
    Oct 2007
    Location
    Tunisia
    Occupation
    Lens Manufacturer
    Posts
    81
    Quote Originally Posted by Paul van Wyk View Post
    We do a calibration on all our lensmeters at least once a year , on the ones in QC we calibrate more often.
    is there a standart procedure to calibrate a lens meter the supplier Of the SHIN NIPPON in our country can not help us !!!!

    Hamza Mahmoud

    Mail : mahmoud.hamza@optylab.com
    Web : www.optylab.com

Thread Information

Users Browsing this Thread

There are currently 1 users browsing this thread. (0 members and 1 guests)

Similar Threads

  1. UV meter
    By frank in forum Optical Marketplace
    Replies: 4
    Last Post: 06-07-2005, 10:22 AM
  2. Pupilo Meter
    By sandeepgoodbole in forum General Optics and Eyecare Discussion Forum
    Replies: 9
    Last Post: 09-27-2003, 11:16 AM
  3. looking for groover and UV meter
    By Jan in forum Optical Marketplace
    Replies: 1
    Last Post: 04-29-2002, 12:07 PM
  4. BPI UV meter for sale
    By rjallison in forum Optical Marketplace
    Replies: 1
    Last Post: 04-07-2001, 03:32 PM

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •