Is it correct that you measure the total near reading power of a freeform progressive lens processed with position of wear to decide if it is good for dispensing?
Announcement
Collapse
No announcement yet.
Freeform Progressive
Collapse
X
-
I do the same as drk, there is a lot of trust in technology necessary with compensated freeform designs, and I trust the process. However, they should still read very close to the compensated power, just make sure you are comparing to the compensated total power, not the prescribed power.
- Likes 1
Comment
-
Elvis you are correct, I misunderstood the OP's question. Freeform progressives should be verified at the distance power circle, as that is the primary power reference of the design. One ought to be able to get a reasonably accurate near power measurement as well, just as with traditional progressives, but that should not be used as the verification power.