    Wide Pds. With High Minus Lenses

    Help!!!! Let me begin by saying that all equipment is properly calibrated and operating well. The equipment used in the process is as follows: National Optronics 3B Blockers, Mr. Blue Blocker, 7E and 7EX edgers, MEI, Marco 101 lensometer and Tomey auto lensometers.

    We have been experiencing wide pds. on high minus lenses (all materials that are -4.00 and up) which call for 3 or more mm. of decentration. No problems with plus lenses. We have been able to compensate for this problem by adding more decentration when blocking with the 3B's or taking off from the Pd when blocking with Mr. Blue. All jobs are edged on geometric center. We do edge all 1.74 on optical center and there is never a problem when blocked on center. For instance a -6.00
    -2.00 which calls for -6 will actually be blocked at -7.5 in and the pds are good. If blocked at -6 in it would definitely be wide. Does this have anything to do with prism and can the prism formula be applied? Is it parallax? Is this caused by the light bending?

    Any advice would be much appreciated.

  2. #2
    Master OptiBoarder
    Join Date
    Feb 2010
    Location
    ALBUQUERQUE
    Occupation
    Optical Laboratory Technician
    Posts
    446
    Since plus power lenses varify and optical center blocked jobs also varify you can eliminate equipment issues. The number that stands out is the fact that jobs are consistantly off 1.5mm which is usually a standard inset value. Check your LMS and make sure correct values are being entered for PD's. Siince plus lenses are not affected perhaps the system is not compensating inset. If the distance PD's are entered in the near pd field the 1.5 is added to achieve a distance value. Layout a test job ..high minus with narrow pd which will calculate -6.0 decentration and varify by manually calculating decentration via measured box method.

  3. #3
    Bad address email on file David@LUXE's Avatar
    Join Date
    May 2013
    Location
    Huntington Beach
    Occupation
    Optical Wholesale Lab (other positions)
    Posts
    9
    Not the LMS. Depending on power and/or amount of decentration compensation could be .5, 1.0, 1.5 or 2.0 additional millimeters. If left uncompensated jobs could be off anywhere from .5 to 2.0 millimeters in each eye. If the LMS was insetting 1.5 mm than the lower powers would be off also. Will check this out first thing tomorrow. I have some help with this tomorrow and will post results. Hard to believe this thread has had nearly 200 views and only one response. Thanks for you help Precisionlab

  4. #4
    Compulsive Truthteller OptiBoard Gold Supporter Uncle Fester's Avatar
    Join Date
    Dec 2004
    Location
    At a position without dimension...
    Occupation
    Dispensing Optician
    Posts
    5,309
    Quote Originally Posted by David@LUXE View Post
    Hard to believe this thread has had nearly 200 views and only one response. Thanks for you help Precisionlab
    Your headline makes all of us look while you really need a headline that says something like "National Optronics Edger problem".

    Then those of us who do not deal with edging would ignore it.

    This is also a problem that very few Optiboard posters would be able to help with imo.

    Maybe you can ask a moderator to change the headline?

    Kind regards and I hope Precisionlab answer works!

  5. #5
    Master OptiBoarder OptiBoard Silver Supporter Jubilee's Avatar
    Join Date
    May 2001
    Location
    United States
    Occupation
    Dispensing Optician
    Posts
    3,197
    If this is happening with every combination of blocker and edger then it leaves only a few places to check.

    My first thought is the lensmeter. The higher powers would be more sensitive to slight amounts of prism or pin issues.

    Second is the LMS. Perhaps certain skus have had misinformation or a recent update had some issues.

    Any reason why you aren't blocking OC for all? Are you sure it is not only with one specific blocker or edger?

    Are,you double checking the two lensmeters with each other to verify accuracy? I recall working with a few auto lensmeters that had trouble with higher powers and higher index, though most of that was with power not prism.

    And yes since this is prism you can manipulate Prentice rule to figure if tthe amount of prism is consistent.
    Last edited by Jubilee; 04-06-2015 at 09:11 PM. Reason: additional question

  6. #6
    OptiBoardaholic
    Join Date
    Nov 2003
    Location
    St. Joseph, Michigan
    Occupation
    Dispensing Optician
    Posts
    244
    After going through a similar situation, with a different brand of equipment, we found that it was indeed the pins on the lensometer were the issue. Spot a mid to high power lens in the lensometer; flip it 180 degrees, and spot again. If it's spotting in two different places, it's the pins.
    The bitterness of poor quality remains long after the sweetness of low price is forgotten.

  7. #7
    Master OptiBoarder OptiBoard Silver Supporter Jubilee's Avatar
    Join Date
    May 2001
    Location
    United States
    Occupation
    Dispensing Optician
    Posts
    3,197
    Now talking with a few of the lab guys at the VA, they have had issues with the 3B blockers and their display inducing error on the high minus lenses. The thickness of the len edges would displace the image by small amounts.

  8. #8
    OptiWizard
    Join Date
    Apr 2014
    Location
    California
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    382
    Quote Originally Posted by Jubilee View Post
    Now talking with a few of the lab guys at the VA, they have had issues with the 3B blockers and their display inducing error on the high minus lenses. The thickness of the len edges would displace the image by small amounts.
    easily verifiable by blocking using a pad with no center, then measuring the decentration from the back of the lens. especially if it's more than 1mm :)

  9. #9
    OptiBoard Novice
    Join Date
    May 2014
    Location
    Australia
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    4
    I had the same issue with my Mr.blue combo. Mostly on high minus with smaller PD. I now use stickers and leap pad both sides of the lens and now have no issues. It turns out that the chuck pressure was pushing the blocks to the center of the lens while clamping. It drove me mad for months because the Cyl was always correct, just the PD was off.

    I hope the Aussie terms translate.

    Cheers

    Leighton

Thread Information

Users Browsing this Thread

There are currently 1 users browsing this thread. (0 members and 1 guests)

Similar Threads

  1. Help! High minus, high prism.
    By saskdispenser in forum General Optics and Eyecare Discussion Forum
    Replies: 11
    Last Post: 07-31-2012, 06:11 PM
  2. Need high minus PAL
    By Johns in forum General Optics and Eyecare Discussion Forum
    Replies: 13
    Last Post: 11-10-2006, 11:46 AM
  3. Another Query re High Minus High Index Progressive
    By snoopybird in forum Progressive Lens Discussion Forum
    Replies: 10
    Last Post: 12-01-2005, 08:15 AM
  4. edge coating for high minus lenses??
    By Vicki in forum General Optics and Eyecare Discussion Forum
    Replies: 9
    Last Post: 07-29-2002, 10:02 PM
  5. Oddity of a high minus, high index pal.
    By Cameron in forum General Optics and Eyecare Discussion Forum
    Replies: 2
    Last Post: 07-13-2000, 12:49 PM

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •