    HELP NIDEK 3D-fit LE-900 EX Express Error Code 0209

    Hello forum, did anyone of you ever get this error code 0209 on your NIDEK 3D-fit LE-900 EX Express? I don't have the manual to look it out, the process just stops in the middle and I get that error message. Is there a link I can download the manual online? Any suggestion is appreciated.
    thanks

    optikanimani

  2. #2
    Bad address email on file
    Join Date
    Oct 2009
    Location
    Hawaii
    Occupation
    Dispensing Optician
    Posts
    9
    4.3.12 Error code 0209 (While the front surface of the lens is measured,
    the feelers are positioned beyond its measurable range.) appears.
    Perform the countermeasures described in “4.3.11 Error code 0208 (While the rear surface
    of the lens is measured, the feelers are positioned beyond its measurable range.) appears.”.
    4.3.11 Error code 0208 (While the rear surface of the lens is measured,
    the feelers are positioned beyond its measurable range.) appears.
    Do the feelers (40340-4100) come off?
    Does the rotary encoder (40340-E043) have a
    break?
    Reassemble the feelers (40340-4100) properly.
    (See “6.12”.)
    No
    No
    Yes
    Does the pulse motor (40340-E033) have a
    break?
    Replace the lens measuring ASSY (30430-
    4000). (See “6.13”.)
    No
    Yes
    Does cable CA29 (40340-CA29) have a
    break?
    Replace the lens measuring ASSY (30430-
    4000). (See “6.13”.)
    Yes
    Is the wire (41002-M388) of the lens
    measuring ASSY cut?
    No
    Replace the lens measuring ASSY (30430-
    4000). (See “6.13”.)
    Yes
    Replace the lens measuring ASSY (30430-
    4000). (See “6.13”.)
    Yes
    Clean the gap between the shaft of the feelers
    (40340-4100) and the lug-hole of the shaft.
    No
    Do the feelers (4340-4100) have any defects
    such as bends, etc.?
    Is the symptom improved?
    Replace the feelers (40340-4100). (See “6.12”.)
    No
    Yes
    Perform the calibration of the lens measuring
    ASSY (40340-4000). (See “7.3.1”.)
    No
    Replace the the MAIN board (40347-BA01).
    (See “6.3”.)
    Completion of operation
    Yes
    Hope this helps


  3. #3
    OptiBoard Novice
    Join Date
    Jul 2010
    Location
    Kosovo
    Occupation
    Optometrist
    Posts
    2
    Thanks Tony, but actually the error message is 0206 and it says "Didn't find a lens. Press START SW" after it completes the first step of grinding the lens. The lens is there but I have to press START SW and run the first step one more time then it goes to complete the lens.

Thread Information

Users Browsing this Thread

There are currently 1 users browsing this thread. (0 members and 1 guests)

Similar Threads

  1. HELP ERROR 0501 Nidek 7070
    By kskpkb in forum Ophthalmic Optics
    Replies: 10
    Last Post: 07-16-2021, 06:25 AM
  2. Nidek Le-9000 problem Error code 0206
    By dardan in forum General Optics and Eyecare Discussion Forum
    Replies: 1
    Last Post: 07-23-2010, 12:02 PM
  3. Nidek SE 9090 Express + Nidek ICE 9000 tracer/blocker
    By getnet in forum Optical Marketplace
    Replies: 2
    Last Post: 10-15-2009, 08:39 AM
  4. HELP ERROR 0501 Nidek 7070
    By kskpkb in forum General Optics and Eyecare Discussion Forum
    Replies: 0
    Last Post: 10-10-2009, 01:55 PM
  5. Code Error for Santinelli Edger 7070 3d
    By Drod3210 in forum General Optics and Eyecare Discussion Forum
    Replies: 14
    Last Post: 03-06-2006, 02:29 AM

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •