    Ultra Optics Mini Question

    Hello all,
    With today being a holiday in the states I won't be able to call Ultra Optics till Monday.
    I am having some issues with my Mini. It's been running great for about 3 weeks. ( I bought it used from Optek)
    Today I noticed that when it went to the second chamber to coat, the coating was just barely coming out the tip. So I opened her up, and tried to adjusted the gage to give it a little more juice. Only trouble is, it was on the max setting.
    Any ideas as to why its only dribbling out? and how would I go about fixing that??

    THANKS!!

    G

  2. #2
    Underemployed Genius Jacqui's Avatar
    Join Date
    Oct 2003
    Location
    Frostbite Falls, Mn.
    Occupation
    Optical Wholesale Lab (other positions)
    Posts
    7,417
    Plugged nozzle??

  3. #3
    Bad address email on file
    Join Date
    Jun 2005
    Location
    Sydney, Australia
    Occupation
    Optical Laboratory Technician
    Posts
    78
    What usually happens is either there is a bit of hardened coating in the pipes which causes the block or the filter needs changing. I would change the filter first (dont forget to turn down the coating pump height) and try that. If there is still a little coating coming out, I would say the pipes are blocked.
    If it is, you need to drain all coating liquid and use methanol (or equivalent) and circulate for a long time(15-30mins). ALternatively you may need to replace the pipes, or remove them one by one to check if they are blocked.
    Good luck, let me know if you need more info.

  4. #4
    Rising Star
    Join Date
    Jul 2006
    Location
    Ontario
    Occupation
    Dispensing Optician
    Posts
    69

    Thanks

    Thanks for the suggestions!!

    Turns out it was the pump :p
    So Im fixed and up and running once again!

    G.

  5. #5
    Master OptiBoarder snowmonster's Avatar
    Join Date
    Jul 2005
    Location
    Somewhere
    Occupation
    Optometrist
    Posts
    1,208
    Quote Originally Posted by Websta View Post
    What usually happens is either there is a bit of hardened coating in the pipes which causes the block or the filter needs changing. I would change the filter first (dont forget to turn down the coating pump height) and try that.
    Ha ha. I forgot to turn that back down a few weeks ago and shot coating up to the top of where the HEPA filter would have been. Ours is in a hood, but still. Quite a fountain.
    -Steve

Thread Information

Users Browsing this Thread

There are currently 1 users browsing this thread. (0 members and 1 guests)

Similar Threads

  1. Ultra Optics Mini II available
    By getnet in forum Optical Marketplace
    Replies: 0
    Last Post: 10-10-2007, 08:21 PM
  2. Looking for an Ultra Optics Mini II hardcoater
    By Mirror in forum Optical Marketplace
    Replies: 1
    Last Post: 04-18-2006, 04:42 PM
  3. Want to Buy: Ultra Optics Mini Coater
    By braheem24 in forum Optical Marketplace
    Replies: 0
    Last Post: 08-25-2005, 11:35 PM
  4. Ultra Optics Mini coater
    By Geff Heidbrink in forum Optical Marketplace
    Replies: 2