    PD on wrap frames

    I hear a lot of talk about Rx calculator's for wrap frames, but not much about verifying PD. The common practice of verifying a PD is to take a ruler and measure the distance from the front of the frame. After tracing these frames we have found that the frame PD comes out narrow, so we will take a frame PD prior to blocking and inset the lens more depending on the difference. The PD then verifies correct.

    I am questioning whether this is correct and how does it affect the patient? If you take a PD from the minus side you will get a narrower reading. By doing what we are doing is this creating too narrow of a PD for the patient? So, what is correct? Am I hallucinating here or is this a valid point?

    Bob

  2. #2
    Master OptiBoarder OptiBoard Silver Supporter Barry Santini's Avatar
    Join Date
    May 2000
    Location
    Seaford, NY USA
    Occupation
    Dispensing Optician
    Posts
    6,008
    Quote Originally Posted by Bob Price View Post
    I hear a lot of talk about Rx calculator's for wrap frames, but not much about verifying PD. The common practice of verifying a PD is to take a ruler and measure the distance from the front of the frame. After tracing these frames we have found that the frame PD comes out narrow, so we will take a frame PD prior to blocking and inset the lens more depending on the difference. The PD then verifies correct.

    I am questioning whether this is correct and how does it affect the patient? If you take a PD from the minus side you will get a narrower reading. By doing what we are doing is this creating too narrow of a PD for the patient? So, what is correct? Am I hallucinating here or is this a valid point?

    Bob
    Depending on the degree of wrap angle, the *effective* frame PD will narrow. You can use simple trigonometry to calculate it exactly. Just visualize that the wrapping of a frame would effectively tighten it against a client's temples, and you'll undertand conceptually how the finished frame PD becomes more narrow

    Barry

  3. #3
    OptiBoardaholic bt5050's Avatar
    Join Date
    Feb 2006
    Location
    USA
    Occupation
    Dispensing Optician
    Posts
    272

    wrap frames - pd

    I have found that when i get jobs in from zeiss - and they use their spazio software - the pds - does seem to come out exact - with the base in prism they induce - as well as the adjustements made to the rx-

    do have a few ?'s -

    1- when you are finaling a rx - that uses a adjusted pd - and rx -
    do you apply the adjusted PD + rx to apply the tolerances - or do you the actual perscribed rx- and pd ??

    2- is it correct to measure these pd's from the back side - do u get a more actuarate measurment - since you are not compensating or reading the ruler at such an angle - ?

    3- when laying out this work - to be edged - should you also - take the frame pd - from the backside - and use that for your decentration- or should you - take the a + the dbl - and then calulate the decentration on these wrap frames ?

    look forward to your answers -

    b

  4. #4
    Master OptiBoarder lensgrinder's Avatar
    Join Date
    Jul 2005
    Location
    Raleigh, NC
    Occupation
    Lens Manufacturer
    Posts
    504
    The shortest distance between two points is a straight line.
    When taking a Frame PD the key is not to bend your PD stick, if this requires you to take the measurement from the back (this method seems to be problematic due to the temples, sometimes), take DBL + A, use Frames Book measurements, then do so.
    The patient's PD is the same, do not bend your PD stick, take it from the back if needed.

    If you are using a compensated Rx, then you would apply ANSI to the compensated Rx, because this is what they are made to, not the original Rx.

  5. #5
    OptiBoardaholic
    Join Date
    Feb 2006
    Location
    Indonesia
    Occupation
    Optical Laboratory Technician
    Posts
    250

    PD on wrap frames

    Quote Originally Posted by Barry Santini View Post
    Depending on the degree of wrap angle, the *effective* frame PD will narrow. You can use simple trigonometry to calculate it exactly. Just visualize that the wrapping of a frame would effectively tighten it against a client's temples, and you'll undertand conceptually how the finished frame PD becomes more narrow

    Barry
    Is there any calculator to determine the PD on wrap frames? Thanks.

    Samuel

  6. #6
    Vision Equipment OptiBoard Corporate Sponsor Leo Hadley Jr's Avatar
    Join Date
    Mar 2003
    Location
    Florida
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    1,664
    This is a great subject,
    not only for the compensated PD, but the RX as well. I cant tell you how many times I have seen older licensed opticians rejecting these jobs because they do not believe compensations have to be made for the wrap. I think it is very important this subject be covered extensively in continuing education, trade magazines, and any other resource like Optiboard for the non licensed states.
    I believe Harry Chiling has a tilt and wrap calculator posted in the file section of OptiBoard.
    That may help out a great deal.
    Leo Hadley Jr
    Vision Equipment
    T: 855.776.2020

    www.visionequipmentinc.com

  7. #7
    Allen Weatherby
    Join Date
    Jun 2004
    Location
    Florida
    Occupation
    Other Optical Manufacturer or Vendor
    Posts
    1,286

    Wrap frames and PD

    Quote Originally Posted by Bob Price View Post
    I hear a lot of talk about Rx calculator's for wrap frames, but not much about verifying PD. The common practice of verifying a PD is to take a ruler and measure the distance from the front of the frame. After tracing these frames we have found that the frame PD comes out narrow, so we will take a frame PD prior to blocking and inset the lens more depending on the difference. The PD then verifies correct.

    I am questioning whether this is correct and how does it affect the patient? If you take a PD from the minus side you will get a narrower reading. By doing what we are doing is this creating too narrow of a PD for the patient? So, what is correct? Am I hallucinating here or is this a valid point?

    Bob
    Consider the a spherical lens with a mono PD of 30mm. With a high wrap frame you may get a needed PD for lens mounting of about 31.5mm. This will place the optical center in front of the pupil. The object of this compensation would be to place the lens in front of the pupil and to do this you need to increase the PD during lens fitting. If you measure a wraped frame link this you would find the distance, when measured on the bench, from the center of the frame to the center of the lens would be 31.5mm. When placed on the patient and then measured keeping the PD stick on the same plane as traditional flat dress eyewear you would measure the PD as 30.0mm.

    There are a lot more calculations that go into mounting lenses in wrap frames, but this should help you understand the basics of the PD.

    Base in prism accomplishes the same thing as increasing the PD.

  8. #8
    OptiBoardaholic bt5050's Avatar
    Join Date
    Feb 2006
    Location
    USA
    Occupation
    Dispensing Optician
    Posts
    272
    i think these explanations ( the photos where helpful for me ) - that r found on the ziess spazio site -
    it shows how not compensating the rx- or pd - alters the actual rx - and the pupil- oc relationship changes -

    http://www.solatechnologies.com/prod...optics/spazio/

    at what power - should compensation start being used ? - it appears they may use - 3d- but i am unclear if this is just their example - or that is what they suggest

    at 1st - i was a bit concerned using -these altered rx and induced prism - jobs - and it came from my exp. that a percribing MD and od - todl the pts -the glasses where made way wrong - and was not happy about it -

    although i got a little satisfaction - by calling him - and showing him how wraps are done - and that we where actually correct - - it is not always conveyed to the the pts very well - and all they hear is - they made them wrong -

    so now i make it a point to mention - at the time of sale - that the rx- will be compensated for the wrap - so they are actually made to the correct rx- just in case in the small % - they go back for a recheck - without seeing me 1st - and get misinformation

    have a great weekend -
    b
    Last edited by bt5050; 04-27-2007 at 08:25 AM. Reason: spelling - link

  9. #9
    ATO Member HarryChiling's Avatar
    Join Date
    Apr 2005
    Location
    Nowhereville
    Occupation
    Other Eyecare-Related Field
    Posts
    7,765
    I have suggested in te past that when doing a wrap the lenses be engraved or laser marked with something to give other a way of reversing the Rx, for example if the lens had engraving T15W20 an optician can determine that the Rx was compensated for 15o tilt and 20o wrap compensation. Then if the formulas used to determine the compensations were reversed (simple as the compensations would be multiplying instead of dividing into the power matrix the compensations) then the patient upon dispense is given a small business card size CD with a compensation program on it (example: Darryl's or mine) wich also has a reverse function so that anyone down the line can reverse the compensations and see what the original Rx was. This would add great value to a lens package and cost very little to the manufacturer.

    If any manufacturer applies any of these techniques to their lenses and comes up with some ultra premium product as a thank you please give me a hefty discount on it.
    1st* HTML5 Tracer Software
    1st Mac Compatible Tracer Software
    1st Linux Compatible Tracer Software

    *Dave at OptiVision has a web based tracer integration package that's awesome.

  10. #10
    ATO Member HarryChiling's Avatar
    Join Date
    Apr 2005
    Location
    Nowhereville
    Occupation
    Other Eyecare-Related Field
    Posts
    7,765
    To find the Rx that should be surfaced into the lens use this calculator:
    http://www.optiboard.com/forums/show...43&postcount=1

    Once the Rx is figured out Use this calculator to come up with you surfaceing ticket:
    http://www.optiboard.com/forums/show...16&postcount=1

    Now with the back curves figured out use this excell sheet to find out what second cuts (lenticular) to make to thin out the edges:
    http://www.optiboard.com/forums/show...04&postcount=3

    I will eventually incorporate all these calculators into one huge wrap and tilt calculator that will blockout the sun and bring about another ice age, but time is a prescious commodity right now. ;)
    1st* HTML5 Tracer Software
    1st Mac Compatible Tracer Software
    1st Linux Compatible Tracer Software

    *Dave at OptiVision has a web based tracer integration package that's awesome.

  11. #11
    Master OptiBoarder lensgrinder's Avatar
    Join Date
    Jul 2005
    Location
    Raleigh, NC
    Occupation
    Lens Manufacturer
    Posts
    504
    Quote Originally Posted by AWTECH View Post

    Base in prism accomplishes the same thing as increasing the PD.
    Please correct me if I am wrong, but the prism is due to the lens tilt, thickness of the lens, and base curve. If you increase the PD, the angle still remains the same. The person is never going to look over the OC, because the light is coming in at an angle to the line of sight, not straight. This is why you need the prism, to correct for the displacement and based on a 14mm vertex and 15 degrees of wrap you would need to increase the PD about 7mm per eye. Let me know your thoughts.

  12. #12
    OptiBoardaholic bt5050's Avatar
    Join Date
    Feb 2006
    Location
    USA
    Occupation
    Dispensing Optician
    Posts
    272

    cd

    then the patient upon dispense is given a small business card size CD with a compensation program on it

    i think your idea with a engraving and something for the pt at disp ( ie card) - would be great - so it is apparent when the rx has been changed from the orginal - provided by the persriber -

    however - i dothink going as far as giving a pt a cd- is kinda over board - ( at least here where i am ) - since the od.md's expire there rx's - the pt would not have the rx for years and years - and need the rx converted back -

    furthermore - if we wanted to make another pr - it is most likely it will be in a differant frame - with differant wrap- and therefore we could easily work off the real rx -


    BUT I DO LOVE THE ENGRAVINGS - and dipense card - idea -!!!

    b

  13. #13
    ATO Member HarryChiling's Avatar
    Join Date
    Apr 2005
    Location
    Nowhereville
    Occupation
    Other Eyecare-Related Field
    Posts
    7,765
    Please correct me if I am wrong, but the prism is due to the lens tilt, thickness of the lens, and base curve. If you increase the PD, the angle still remains the same. The person is never going to look over the OC, because the light is coming in at an angle to the line of sight, not straight. This is why you need the prism, to correct for the displacement and based on a 14mm vertex and 15 degrees of wrap you would need to increase the PD about 7mm per eye. Let me know your thoughts.
    The prism is for alighning the ray when leaving the lens to be parralel with the optical axis, the compensation to the PD would be for the ray from lens front through the lens IMO. So if we were to consider this the compensation would be:

    tan(wrap)=dec/thickness

    for example a lens 2mm thick wraped 15o, would be:

    tan(15)=dec/2mm
    tan(15)*2mm=dec
    dec=tan(15)*2mm
    dec=0.2679*2
    dec=0.5358

    So add 0.5mm to your lens PD.

    Quote Originally Posted by bt5050
    however - i dothink going as far as giving a pt a cd- is kinda over board - ( at least here where i am ) - since the od.md's expire there rx's - the pt would not have the rx for years and years - and need the rx converted back -
    The idea is so that the prescribing doctor can reverse the Rx themselves and would have the software for further use. It would be more to add value than anything else. You could hand out a card with the Prescribed Rx and the Compensated Rx, but if the card is lost the glasse are wrong again. However if the engravings catch on then when a preescriber sees it they will know to use the software to get the original Rx. The patient doesn't have to worry about anything and the prescriber now has a tool at their disposal whihc would further prommote their product. Just an idea, I can dream can't I. :D
    1st* HTML5 Tracer Software
    1st Mac Compatible Tracer Software
    1st Linux Compatible Tracer Software

    *Dave at OptiVision has a web based tracer integration package that's awesome.

  14. #14
    Master OptiBoarder Darryl Meister's Avatar
    Join Date
    May 2000
    Location
    Kansas City, Kansas, United States
    Occupation
    Lens Manufacturer
    Posts
    3,700
    The prism is for alighning the ray when leaving the lens to be parralel with the optical axis, the compensation to the PD would be for the ray from lens front through the lens IMO. So if we were to consider this the compensation would be
    This only applies to prism induced by decentration. The tilt of the actual lens, with or without any decentration, also induces a prismatic effect that is independent of decentration and proportional to the front curve of the lens. Moreover, the tilt of a lens is further exacerbated by decentration, even when the PD is correct, since the decentration of a meniscus lens occurs about an arc. While I provide a fairly accurate first order approximation of these effects in my Rx Compensator program, my impression is that very few "wrap calculators" take these variables into account.

    And this all assumes that the PD was correct in the first place. Since wrap tilt effectively reduces the optical center distance, by an amount roughly proportional to the cosine of the angle of tilt, this needs to be factored in before edging the lenses -- either in the compensation software or by the person laying out the lens for edging.
    Darryl J. Meister, ABOM

  15. #15
    Master OptiBoarder OptiBoard Silver Supporter Barry Santini's Avatar
    Join Date
    May 2000
    Location
    Seaford, NY USA
    Occupation
    Dispensing Optician
    Posts
    6,008
    Quote Originally Posted by Darryl Meister View Post
    And this all assumes that the PD was correct in the first place.
    Yes, we all need to review whether our traditional methods for "taking a PD" are, in fact...*representative* of the actual visual axis...

    Since wrap tilt effectively reduces the optical center distance, by an amount roughly proportional to the cosine of the angle of tilt, this needs to be factored in before edging the lenses -- either in the compensation software or by the person laying out the lens for edging.
    This type of PD compensation is already on my wish list for Darryl's next version of his Rx compensator...

    How 'bout it, Darryl????!


    barry

    me

  16. #16
    Master OptiBoarder Darryl Meister's Avatar
    Join Date
    May 2000
    Location
    Kansas City, Kansas, United States
    Occupation
    Lens Manufacturer
    Posts
    3,700
    This type of PD compensation is already on my wish list for Darryl's next version of his Rx compensator... How 'bout it, Darryl????!
    I'll get it in there sooner or later... I just have to figure out the best way not to confuse people with it.
    Darryl J. Meister, ABOM

  17. #17
    OptiBoardaholic
    Join Date
    Feb 2006
    Location
    Indonesia
    Occupation
    Optical Laboratory Technician
    Posts
    250
    Darryl, is this kind of lens what we called Loop?

  18. #18
    OptiBoardaholic bt5050's Avatar
    Join Date
    Feb 2006
    Location
    USA
    Occupation
    Dispensing Optician
    Posts
    272

    trade mag

    hi everyone -

    have been looking around for a old trade mag. that i saw - that had a great write up on wrap frames - and it had great explanations on the optics outcome - and the ways these issues are handled -

    i was looking for this - to provide to my staff - to have them get a better understanding of what is going on - and why -

    i thought the pics - where great also - for those that need to see it - to understand it -

    any help would be appreciated -
    b

  19. #19
    ATO Member HarryChiling's Avatar
    Join Date
    Apr 2005
    Location
    Nowhereville
    Occupation
    Other Eyecare-Related Field
    Posts
    7,765
    I think it might have been a 20/20 magazine, I will look through my office for it I am almost sure I have an extra copy.
    1st* HTML5 Tracer Software
    1st Mac Compatible Tracer Software
    1st Linux Compatible Tracer Software

    *Dave at OptiVision has a web based tracer integration package that's awesome.

  20. #20
    Master OptiBoarder
    Join Date
    Oct 2004
    Location
    Down in a hole!
    Occupation
    Dispensing Optician
    Posts
    13,078

    20/20 Wrap Article

    I think a fellow Optiboarder=Barry Santini, wrote that one.

  21. #21
    Master OptiBoarder OptiBoard Silver Supporter Barry Santini's Avatar
    Join Date
    May 2000
    Location
    Seaford, NY USA
    Occupation
    Dispensing Optician
    Posts
    6,008
    Quote Originally Posted by Fezz View Post
    I think a fellow Optiboarder=Barry Santini, wrote that one.
    Sorry, Fezz, as much as I'd like to take credit for it...I didn't write that one. That was Dave Carr, and you can see it on the Jobson website for free.

    Check it out: http://www.2020mag.com/CE/TabViewTes...6/Default.aspx

  22. #22
    Master OptiBoarder
    Join Date
    Oct 2004
    Location
    Down in a hole!
    Occupation
    Dispensing Optician
    Posts
    13,078

    I goofed---again!

    OOPPPPSSSS!

    I was thinking of your "Get Going With Goggles" January 2007 20/20 issue.

    My bad!

  23. #23
    Master OptiBoarder OptiBoard Gold Supporter DragonLensmanWV's Avatar
    Join Date
    Oct 2006
    Location
    The Greatest Nation
    Occupation
    Optical Retail
    Posts
    7,645
    Quote Originally Posted by lensgrinder View Post
    The shortest distance between two points is a straight line.
    When taking a Frame PD the key is not to bend your PD stick, if this requires you to take the measurement from the back (this method seems to be problematic due to the temples, sometimes), take DBL + A, use Frames Book measurements, then do so.
    The patient's PD is the same, do not bend your PD stick, take it from the back if needed.

    If you are using a compensated Rx, then you would apply ANSI to the compensated Rx, because this is what they are made to, not the original Rx.
    That's a good argument for the wooden PD rulers that I use.
    DragonlensmanWV N.A.O.L.
    "There is nothing patriotic about hating your government or pretending you can hate your government but love your country."

  24. #24
    What's up? drk's Avatar
    Join Date
    Mar 2004
    Location
    Ohio
    Occupation
    Optometrist
    Posts
    9,388
    A corollary that has nothing to do with this subject, directly:

    Picture this cranial geometry with pupillometry, from an overhead view:

    Coronal plane of left orbit relatively planar to coronal plane of bridge, but coronal plane of right orbit curved away from coronal plane of bridge (say, from a front-on view, "flat face" from left eye to bridge and "rounder face" from bridge to right eye.) Would not the pupillometer reading be automatically correct? The vertex distance from right eye to pupillometer will be greater than the left, and the p.d will be read "too short" at the linear plane of the pupillometer, but the glasses will be worn in the same plane as the pupillometer, anyway, so all works out.

    But, what if the silver top bar of the pupillometer is not "up" or up but not touching the forehead to make the pupillometer approximate the plane of the glasses frame, and the pupillometer "rolls" towards the right via the rounder forehead...will that not introduce somewhat significant error in the p.d. measurement? It would OVERESTIMATE the measurement on the eye that it was pushed too far towards, and UNDERESTIMATE the measurement on the eye that it was tilted away from.

    Likewise, if the pupillometer measurement were done exactly right with the top bar touching, if the frame at dispense were not adjusted the exact same way, say with the vertex distance incorrectly short on one side from non-parallel temple splay, think of how that would effect the vision on narrow progressive corridors and reading areas! It would wreak havoc.





    *******
    For that matter, if an ophthalmic frame has more wrap (by design or by optician) than the flat pupillometer, the lenses will have the p.d.s effectively too narrow AND THE PROGRESSIVES INSET TOO FAR.

    If the ophthalmic frame has less wrap than a flat pupillometer (think slightly negative face form on a fat head with a small frame with a memory metal bridge), the lenses will have p.d.s effectively too wide AND THE PROGRESSIVE OUTSET TOO FAR.

    (To make this germane to the thread, consider wrapped sun progressives! What a source of multiple errors galore! It's suprising anyone can wear those without a physics software program! Not only would there be mind-numbing VARIABLE tilt effects from top to bottom, but the pd on those progressives are critical, and the frame wrap p.d. effect would about HAVE to be compensated for, in order to get binocular vision at near. Why do I even try to do it? It seems like sheer folly! FT-28 set low, here I come.)
    Last edited by drk; 05-13-2007 at 12:00 AM.

  25. #25
    Master OptiBoarder OptiBoard Silver Supporter Barry Santini's Avatar
    Join Date
    May 2000
    Location
    Seaford, NY USA
    Occupation
    Dispensing Optician
    Posts
    6,008

    DRK says *EXACTLY* what I'm thinking...

    Pupilometers make certain assumptions during fitting that are not always representative. And I was lookin at the Zeiss *EYE TERMINAL* as a possible solution, but I'm thinkin' how would one *verify* the finished eyewear, in situ, with the Eye terminal? Revert to felt tip pen? Well, then what's is the point of Eye Terminal?

    My two cents, what's yours?

    Barry

Thread Information

Users Browsing this Thread

There are currently 1 users browsing this thread. (0 members and 1 guests)

Similar Threads

  1. Wrap Frames W/ Transitions Bifocals
    By Steelerfan in forum General Optics and Eyecare Discussion Forum
    Replies: 2
    Last Post: 04-05-2007, 03:25 PM
  2. wrap distortion
    By rayban in forum General Optics and Eyecare Discussion Forum
    Replies: 12
    Last Post: 10-13-2006, 05:21 PM
  3. Wrap Safety frames
    By mattress in forum General Optics and Eyecare Discussion Forum
    Replies: 7
    Last Post: 02-20-2006, 05:32 PM
  4. Calculation of edge thickness in wrap frames.
    By Jedi in forum Ophthalmic Optics
    Replies: 6
    Last Post: 09-23-2005, 09:29 AM
  5. wrap frames and RX - Is there a limit?
    By Oriki in forum General Optics and Eyecare Discussion Forum
    Replies: 11
    Last Post: 07-08-2005, 06:11 PM

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •